Waters HPLC problem
Hi We purchased a refurbished WATERS HPLC, including 1525 pumps and 2487 detector. First we found both pump and detector were not connected to PC when using Empower. We figured out the reason was the broken BUS/LACE card. We recently replaced BUS/LACE card, but now the problem is, 2487 detector showed connected, but 1525 pump is not. That is strange. Does anybody know why? Is there any broken part inside pump to cause this?
Thanks a lot
Thanks a lot
0
Answers
-
IEEE address to be set up in the pump. You have to check for the dip switches in the pump backside. If there are modules connected and have IEEE address conflict, you can have this situation. Also the IEEE cable used for the pump may be malfunctioning. Use the cable connected to your detector with pump and confirm whether pump is communicating0
-
blr81 said:IEEE address to be set up in the pump. You have to check for the dip switches in the pump backside. If there are modules connected and have IEEE address conflict, you can have this situation. Also the IEEE cable used for the pump may be malfunctioning. Use the cable connected to your detector with pump and confirm whether pump is communicating0
-
If each instrument has dip switches they all must be set unique. The Empower software cannot communicate until each is manually set. There should be instructions to set these included with the BL card. Try Steaven's suggestion and just connect one at a time. You may also need to start over in the Empower Node and System setup fields. Delete both, restart all, reboot lace. Good luck, BL is normally very easy!0
-
JCarter280 said:If each instrument has dip switches they all must be set unique. The Empower software cannot communicate until each is manually set. There should be instructions to set these included with the BL card. Try Steaven's suggestion and just connect one at a time. You may also need to start over in the Empower Node and System setup fields. Delete both, restart all, reboot lace. Good luck!0
-
Im not sure about 1525's... but I thought those single solvent high pressure pumps did not have IEE control directly, but though a special Pump Control Module.