Hello Phillip,
You should not depend on this backup until the cause can be determined. What entries do you see in the Audit Trail at the time of the backup? There should be an entry or multiple entries relating to and stating the backup failed.
Also you mentioned other entries in the Message Center. Can you be specific about these entries? They could provide good information.
Mapping errors usually relate to physical or virtual memory. Have you ruled out if the issue is client specific?
Look forward to hearing from you soon,
Dot
0 -
The Audit Trail doesn't show anything out of the ordinary: Backup started, cancelled, or finished; Project Integrity passed or failed. The other messages in the Message Center were simply the same message for each data file affected, a handful in each project. It seems to be the same on a local disk or the network drive, both. Here, you can see what I mean:
? Type Category Time Application Message 67 Error General 4/19/2011 12:58:51 PM EDT Configuration Manager MapFileForRead file C:Backupd44562.dat failed 68 Error General 4/19/2011 12:58:51 PM EDT Configuration Manager MapFileForRead file C:Backupd44565.dat failed 69 Error General 4/19/2011 12:58:51 PM EDT Configuration Manager MapFileForRead file C:Backupd44568.dat failed 70 Error General</... 0 -
Hello Phillip,
The Message Center indicates some raw datafiles are not present. If you would, verify the files d2681.dat, d2753.dat, and d2764.dat are present in the folder EmpowerProjectsDrug. If they are present what size are the files?
Thank you
Dot
0 -
Niether d2681, d2753, nor d2764.dat exist, but d2682. d2754, and d2765.dat do exist. They are 5, 2, and 2kb respectively.
0 -
Hello Phillip,
This issue requires more investigation that is not easy in this forum.
If it is agreeable, I will be in touch to continue working with you.