Degasser vacuum HW error
Answers
-
Hello
Thi error literally is related to the deagser. I think if you have checked for the obvious fluidic problems, but just check the lines from the degasser seem tight and there are no obvious leaks. AFter that you have a degasser chamber failure or this this is an electronics issue. I think this will need a ervice call to be made to your local ervie representative.
Liz
0 -
How did you resolve the problem? I just have same problem with you. Thanks!
0 -
It required a service visit to replace the degasser unit.
0 -
Thanks!
0 -
This sounds like a weak vac. pump on degasser. The degasser wants to see the vac. reach a certain level and maintain it. These pumps wear out with age. Check degasser vac. in console. Ideally you should see a reading of .9 or less. Also these pumps often begin to make a faint clunking sound as the bearings start to go.
If you do repairs yourself pump not very difficult to change out. Part # 700001352.
0 -
I had this issue today. I found a hint somewhere that indicated powering off and on might help.I cycled power on and off a several times and the eventually the problem cleared.0
-
Note - while it is a good idea to make sure that the degasser pump is the culprit before just replacing it, know that you can order the part and replace it yourself in about 5 minutes (assuming you're not a menace when touching tools). It's the same pump for 2695s as for Acquity systems.0