It appears that the SM cannot attain any better than 9 degrees in your environment. Efficiency does depend on ambient temperature and humidity, high ambient (above 21 degrees) and/or high humidity (above 50%) may be the issue. Also ensure that you have the defrosted recently. I will attach the specification.
This may be part of the issue, but I also believe that you have run into the way the temperature error band works and we have addressed this in in v1.40 driver. In the method window you set a temperature and an error band. For example if it is 10 degrees and the error band is 5 degrees, the software interprets this as +/- 5 degrees, in other words it will not start to run until under 12.5 degrees and/or throw an error message if the threshold is crossed. You can simply disable the error band and that will allow you to run.
Hope this helps
Liz
0 -
thanks for help Liz