Re: Sample pressure errors
Here is a list of ways to avoid SAMPLE MANAGER LOW PRESSURE ERRORS During the transfer of sample from the sample needle to the sample loop, the Sample Manager pressurizes the flow path to 100 PSI. If…1 -
Re: Is there any way to see the result of an intersample custom field without processing sample set?
it would be a nice addition to the software but i think it is to complicated to make it work. The review of injections would get superslow cuss it always has to recalculate all injections to search f…1 -
Re: How can I create a report where the SST-table shows up before each sample-result.
Another way to do this is to put your SST table in the header and have it on every page?1 -
Re: PATROL UPLC Process Analysis
Hi Doria, here are some things I would look at: 1.) Dynamic Leak Test Failing - you mentioned you replaced Check Valves, I'm assuming Engineer Replaced the Plunger Seals as well? If not this would be…1 -
Re: RRT Impurity: change decimals to 2
Thank you.1 -
Re: Injector Linearity and Accuracy Test (SQT + Supplemantary Test)
Hi MarcNoble, I observed this 2 times and only with 10uL nominal loop volume. The injector linearity with 5 and 20uL loop also passed, no problem with them. But maybe it's just coincidence or as you …1 -
Re: Accounting for system and solvent peaks when reporting impurites?
Go back to your first CF and add the condition there: (TimeCorr*Area*LTD*60)+EQ(Peak Label,"Blank")*-1*50000 and make sure to tick All or Nothing in the formula editor or else it wont work.…1 -
Re: What is this doing in my Custom Field?
Thank you for your answer. I appreciate your help.1 -
Re: Custom Field subtracting average area of "blank"
I was able to get this to work using the peak enumerated CF with the Use As Field selected. I did stumble upon this before you answered on Friday but it wasn't quite working. It wasn't until I saw in…1 -
Re: Custom Field subtracting average area of "blank"
Can you try using the "Summarize Custom Fields Incrementally" function and insert it across various lines of the sample set? So if you have the first sequence of EXS1, followed by S1, then …1