Random Instrument failure with H-Class

Facing random"instrument failure" issue with H-Class system firmware 1.6,Empower 3  SR2 FR2
Module connected with QSM,SMFTN,30CM-CHA, PDA and Sample Organizer.
So far replaced LAC box , Ethernet card & Cables, FTN control cable, 30CM CH-A PCB.
No error message in console. "instrument failure" message appears in Empower message centre.
Reloaded ICS, Created new project & system. LAN power saving setting is OFF. Firewall OFF.
Instrument fails when it goes for injection.
Can someone help here 

Welcome!

It looks like you're new here. Sign in or register to get started.

Answers

  • I recall facing a similar issue a while back. Waters had to come in and replace some parts in the UPLC. I don't remember the specifics. You pretty much narrowed it down to the instrument by replacing the LACE and cables.
  • Please check the Acquity ICOP is above DP4 SR1 or the latest one DP2017 R1 installed .
    Secondly ensure the network speed is good from Lac/e to the Servers and no packet drops.
  • DP2017 R1  will resolved all these issues, make sure instrument firmware also get updated.