Low Resolution on Waters Acquity Peptide CSH C18 150mm columns
Hi,
I've been having some problems with resolution between peaks using Waters Acquity Peptide CSH C18 columns for routine system post-maintenance checks. We use a peptide standard mix which gives 5 peaks, and the resolution between 2 of the peaks is consistently low and not meeting acceptance criteria.
I've repaired damaged PEEK tubing at the column outlet end and washed the column in 10:90 0.1% TFA in water: 0.1% TFA in ACN overnight, with no improvement. Apart from ordering a new column, is there anything else I should try?
Thank you
I've been having some problems with resolution between peaks using Waters Acquity Peptide CSH C18 columns for routine system post-maintenance checks. We use a peptide standard mix which gives 5 peaks, and the resolution between 2 of the peaks is consistently low and not meeting acceptance criteria.
I've repaired damaged PEEK tubing at the column outlet end and washed the column in 10:90 0.1% TFA in water: 0.1% TFA in ACN overnight, with no improvement. Apart from ordering a new column, is there anything else I should try?
Thank you