three different calibration curves in one result set
Hi,
I need to calculate the amount of three Samples with three different calibration curve.
The sequence is formed by three injection of sample and three injection of STD A alternated.
To calculate the amount of the three injections of Sample 1 I need to create a calibration curve with the first three STD A. Then, to calculate the amount of the Sample 2, I need a new calibration curve made up of the next three standards. Same for Sample 3
Sample 1a
STD A
Sample 1b
STD A
Sample 1c
STD A
STD B
Sample 2a
STD A
Sample 2b
STD A
Sample 2c
STD A
STD B
Sample 3a
STD A
Sample 3b
STD A
Sample 3c
STD A
STD B
I tried in this way but it doesn't work. It considers the STD incrementally in the calibration curve.
Can someone help me?
I need to calculate the amount of three Samples with three different calibration curve.
The sequence is formed by three injection of sample and three injection of STD A alternated.
To calculate the amount of the three injections of Sample 1 I need to create a calibration curve with the first three STD A. Then, to calculate the amount of the Sample 2, I need a new calibration curve made up of the next three standards. Same for Sample 3
Sample 1a
STD A
Sample 1b
STD A
Sample 1c
STD A
STD B
Sample 2a
STD A
Sample 2b
STD A
Sample 2c
STD A
STD B
Sample 3a
STD A
Sample 3b
STD A
Sample 3c
STD A
STD B
I tried in this way but it doesn't work. It considers the STD incrementally in the calibration curve.
Can someone help me?
0
Best Answer
-
Set the processing to "don't process (or report)" for all your injections.
Only set normal for your Calibrate,Quantitate, Clear Calibration and summarize custom fields,
Than it should work.
1
Answers
-
it works! Thanks