three different calibration curves in one result set

Zonghi
edited July 11 in Software Products
Hi, 
I need to calculate the amount of three Samples with three different calibration curve. 
The sequence is formed by three injection of sample and three injection of STD A alternated.
To calculate the amount of the three injections of Sample 1 I need to create a calibration curve with the first three STD A. Then, to calculate the amount of the Sample 2, I need a new calibration curve  made up of the next three standards. Same for Sample 3

Sample 1a
STD A
Sample 1b
STD A
Sample 1c
STD A
STD B
Sample 2a
STD A
Sample 2b
STD A
Sample 2c
STD A
STD B
Sample 3a
STD A
Sample 3b
STD A
Sample 3c
STD A
STD B



I tried in this way but it doesn't work. It considers the STD incrementally in the calibration curve. 

Can someone help me?

Best Answer

  • Jovvor
    Answer ✓
    Set the processing to "don't process (or report)" for all your injections.
    Only set normal for your Calibrate,Quantitate, Clear Calibration and summarize custom fields, 

    Than it should work.