Default calculations in Empower 3

skamppanan
edited February 2021 in Software Products

Review on the data report we have areas, USP resolution, Plate count

Do I need to repeat verification between the area display in review versus my report. because by my opinion this should be the same if the software is validated (21CFR part 11) 

Comments

  • That depends on what you're reviewing.
    If you're reviewing channels, all bets are off because you're looking at raw data filtered through a processing method.

    Once you're actually processed the channels with a processing method and you're reviewing results (instead of channels), things are locked down. 

    For results, columns in your review window will match what's in the report (unless there are some custom fields that have not been summarized or something).

    Any other problems can usually be traced back to sample set entries that are either not there or are wrong.