Best Answer
-
Cant you incorporate the purity into your Value? By working out the concentration already? Say you weigh out 10mg into 50mls of a 110% pure standard, which would be a (10/50)*(110/100) = 0.22 mcg/mL, this goes into Value and bypasses Purity. Only Empower 3 FR4 onwards allows the use of values up to 150% in the Purity Tab.
Failing that, I cant see another way for this except as you mention and put the purity in the Samplweight column of the standard and 100 in the dilution column. That only works if all components in a standard mixture share the same purity value, too.
0
Answers
-
Empower2019 said:
Cant you incorporate the purity into your Value? By working out the concentration already? Say you weigh out 10mg into 50mls of a 110% pure standard, which would be a (10/50)*(110/100) = 0.22 mcg/mL, this goes into Value and bypasses Purity. Only Empower 3 FR4 onwards allows the use of values up to 150% in the Purity Tab.
Failing that, I cant see another way for this except as you mention and put the purity in the Samplweight column of the standard and 100 in the dilution column. That only works if all components in a standard mixture share the same purity value, too.
That said, I guess I won't have a problem as soon as we migrate to FR 5. I missed that they had changed the purity field properties.