Table for UV maxima in the report method
Dear all,
In a project of us we must check the identity of the API. According to our SOP the both UV maxima must not deviate by more than 1 nm from the value given in the SOP. Is it possible to create a table in the report method that lists all UV maxima of the corresponding peak?
I have found the Field "Peak Lambda max." that shows the UV maximum. But the substance has two UV maxima...
Thanks in advance!
0
Answers
-
I believe under the 3D /PDA section in report publisher there is an option called Extraction table which lists all the wavelength and AU values across the chromatogram.you can set a min and max wavelength range in field boxes.
This may help if you play around with it but i didnt see a retention time or peak name option so it may need editing.0