calibration curve error
Hi ,
Something strange is happening when processing results
2 point calibration curve but intercept is 7.20x10Ttp18
ie typical curve Y= 3.08e+0.05X with R = 1.00000
Before processing results of unknown between standards are correct
When processed the amount results are all the same in that bracket
Is it using an old calibration curve ? .. but only has points see attached
Thanks
Answers
-
sorry messed up image
0 -
Well .... as you are aware that calibration curve is all sorts of not quite right...
Two things that I noticed:
1. No level --- add a level to your calibrant in either your sample set method or processing method; where every it is you put such information.
2. From the looks of your plot, though it has the x-value correct, it seem slike the y-value (peak area) is zero, which is contradicted by the response field in the table. So #1 again, but also check to see if the peak is being identified correctly.
0 -
Hi, increase your decimal places on X Value to 14 to confirm they are identical.0
-
Thanks for reply
Never needed in past to put in levels and worked fine
Indeed y value = 0 (area) but 2 areas clearly shown in calibration table with amount
x values identical to 14 dp = 13.02800000000000
Processing method created from new and wizard seems to have worked and correct calibration curve shown 2.52e+0.05X and not -5.53e+017X+7.20e+018
Suspect data corruption somehow in first bracket when processed
0