FTN: String?
Having issues restarting a dried out ACQ H class. It's throwing degasser hardware faults, presumably because there's no MP in the QSM (yet).
I'm also seeing FTN: String errors.
Are these related??
TIA,
DR
I'm also seeing FTN: String errors.
Are these related??
TIA,
DR
0
Answers
-
Hi Dan,
You may be experiencing two separate issues. The "String" error is a communication issue. Take a look at this article in the Knowledge base:
Your degasser may failing because it is dry The degasser needs the resistance of solvent to achieve vacuum. Try priming for at least 4 minutes per line, then cycle the power.
good luck!
Marc
0