Is it possible to use the MassLynx library functionality with UV-data?
I tried using the library functionality in MassLynx with UV-spectrum data acquired with an Acquity PDA detector, but I keep getting different errors such as "Cannot read spectral data" and "Refine cannot process a continuum datafile". Also the X-axis in the library window is always shown in m/z units.
0
Answers