QSM Problem

Options
The QSM module is not able to connect with LACE.

Currently, The QSM module connected to LACE has “Serial Number Uninitialized” error in the instruments tab under node

Answers

  • 1) On an Empower PC - C:\Empower\Instruments\Firmware\Loader.exe
    2) Select the Scan for instruments box, and then press Next
    3) 
    Right-click the instrument, choose Set Instrument Serial Number, and then specify the SN. 
    4) Click okay from within the set serial number screen to return to the main instrument list