How do you get basline subtration to work in empower 3
Options
<p>Hi,</p><p></p><p>I have run blanks and labled them as blanks and ticked the blank box.</p><p></p><p>What other setting do you need to imput and how can you tell it to just subtract from certian injections in a run ?</p><p></p><p>Reagards</p><p></p><p>Chris cannell</p>
0
Answers
-
Hi Chris,
I don't have Empower 3, so if this functionality is very different from Empower 2, this answer won't be very helpful - sorry about that.
The attached file describes how to do this subraction in Empower 2 with some screen shots. I put this together for an unusual application where you were subtracting the standard from teh sample to look for differences....
Hopefully this can get you going.
Good Luck,
Kevin
0 -
Hi Chris
There is no difference in the procedure on how to do a blank/baseline subtraction in Empower2 and 3 so you can use Kevin´s procedure. The "blank" column you are referring to is used for signal to noise calculations where you can use the noise from the injection labelled "blank".
best regards
Rune
0