Interchannel calculations in Empower?
Answers
-
As far as I am aware, this is the intended functionality and everything is bahaving as normal. If you specify a channel in your custom field, then the field only runs on that channel. If you use the "%" wild card for the channel, all channels will be calculated, but results are based relatively to the channel derived.
I may know where you are going with this though... Let's see if this helps:
I once had to work with transferring a very old and off assay to a UPLC. The old and odd part about the assay was there was a wavelength change mid run. I believe the run started at 270 nm and half way through it switched to 280 nm. Anyway, some components were detected at 270 nm and others at 280 nm; the same applied to system suitability compounds. So, I needed to come up with intersample calculations for both assay results and system suitability at two different wavelengths.
My solution was to use a partial wild card for the channel portion of the custom field, which became "2?0nm" for the channel portion of the custom field. In the processing method, I then "tied" components to a specific detector channel as all components could be detected at both wavelengths. The summary report formatting worked in a smilar fashion due to locking down components to a specific detector channel in the processing method.
I hope that helps you out some...