Peak shape changing; peak heigh gradually decreased to a half
Answers
-
Hi Deleon,
From what I gather, your retention time has shifted on your original instrument and simultaneously your peak height has changed from 0.6 to 0.3 AU. If the the change in peak shape is shows a large increase in tailing factor, I would suspect it is time to change your column. If the change in peak shape is based solely on a broader peak width, it sounds like there may be a problem with the pump pressure. Might want to check the initial pressure for the system as a starting point.
Good Luck,
Brian
0 -
Hi Deleon
Have you changed instrument method parameters ? Could someone accidentally have changed something especially the Detector settings like sampling rate and filter time constant ? Was the change you have seen sudden in the middle of a sample set or was the change observed with a new sample set ?
best regards
0