How do I connect a Waters fraction collector III to my UPLC Acuity system for software control?
Answers
-
Julie,
The WFC III communicates to MassLynx/FractionLynx via RS-232. The part number for the cable is 700002170. I suggest you work through your local Waters Reps to make sure you obtain everything you need.
Regards,
Rich
0 -
Hello Rich,
I'm pretty sure the answer is no but could not resist asking. Is the a 2154 option for WFCIII and ACQUITY? Will there be? Thanks!
Eric
0 -
our existing WFCIII already has the E2 prom, but running with a 2695.
0 -
Eric,
There is no restriction on using the WFC III with an ACQUITY under Empower control. It can and has been done.
That said, considering that the WFC III was designed for use at higher flow rates than those typically encountered in UPLC, you'll need to be creative when working at lower flow rates.
Contact me directly and I'd be happy to discuss this further.
Regards,
Rich
0 -
thank you Rich. greatly appreciate it.
Eric
0 -
Rich, Do I connect it to the detector
via an RS232 or attempt to connect it somehow to the sample manager? Also, what special maneuvers need to be made (to which you referred) in order to work on AcquityUPLC at flows of 0.1 to 0.2ml/min?
thanks, julie
0 -
Hi Julie,
The RS 232 port of the WFC III is connected to an RS 232 port on your MassLynx workstation.
The rest of your questions are probably best answered by telephone.
I'll give you 2 options:
1) Call 800 252 4752 and speak to Technical Support
2) Send me a private e-mail with your contact information and I will call you.
Regards,
Rich
0