PAHs analysis by UPLC- tubing easily got clogged
Comments
-
>> However, I couldn't dissolve my sample in 50/50 ACN/Water at the start because some PAHs may not dissolve well in this solvent. Is there any solution for this? Please advice.
Sounds like the PAHs are crashing out in the tubing, and potentially eating the tubing as the deposits grow. If the analytes won't stay in your mobile phase solution on the bench, they probably won't stay in solution on the LC.
Are you using an H-Class or I-Class system?
If you have a binary pump: mix a liter of 50/50 IPA-ACN, And place that on your solvent line, then on your aqueous line, use whatever you were using before.The mixture should run about 60%-70% Solvent and 30-40% Aqueous. May be worth testing solubility (once at 60-40, and again at 70-30) on the bench first.
If you have a quaternary pump: Start at 30% IPA / 30% ACN / 40% H2O and make adjustments as necessary from there. Again, testing your mobile phase ratios with sample solubility on the bench is a good idea.