Increasing Pressure as the run go
Comments
-
Hi Severine,
I am afraid that this will be tough to resolve here. I have quite a lot of questions...
What exactly are you injecting (and how is it dissolved?) and how much? The pressure increases at each injection and not during the run, so there must be "something" going on there. What are your wash and purge solvents?
What do you do when the pressure has dropped and you get your blank chromatograms? This sounds like a leak in the FTN / Needle Seal... How do get your peaks back?
Are you using In-Line-Filters or Guard Columns in front of your column? Are you filtering the Buffers and/or the samples? What is the buffer concentration and how do you flush the system? Does the borate buffer precipitate when you hit 100% ACN?
I am sorry that I am not answering any questions but only adding so many new ones. Maybe you can answer some of them? This would help a lot.
Best regards
Tobias
0 -
Hi,
I found out why the pressure was increasing steadily before dropping.
The I2V was leaking and the check valve had to be replaced.
Since then the pressure has been perfect.
0