nanoACQUITY problem
Comments
-
Hello,
If you are not doing this already, try making a complete run of the method at least once before making your 'real' injection. One group I was in contact with had a similar problem and came to believe that the system 'healed' itself after several runs. I think the more likely explanation is that the both the trap column and analytcal column, as well as the associated tubing, are left in the correct state of equilibration for the separation to work correctly.
Another reason for losing retention could be that the column has become contaminated with something which decreases its active surface area, effectively making it a shorter column. The only way to test this possibility is to replace it.
Ken Blakeslee
0 -
Hi
Thanks for your recommend. But I found the reason.
I check pump B flow profile. As I thought in 2D with dilution or 2D online mode some solvent B come into column when gradient started randomly.
I think its software problem.
Hangyeore Lee
0