Condensation in Sample Manager
Comments
-
Hi!
We run a method which sample T is 4 deg and we have the same problem. We try to turn off T as soon as we are done but when we can't we just use a lot of paper towels. So far no other problems have been caused becasue of the water, only the extra work. I have always wondered if is a design problem (Water people could you answer that?), sorry if we don't offer a better answer!
0 -
You are my hero! Thank you for confirming it's not just my system!! I agree with you, I think it MUST be a design flaw. Waters??...Anyone care to address this for us??
0 -
Hi
You could check the ICOP version of the Sample Manager to see if it can be updated. Versions older than 1.4 have a different defrost cycle for the peltier device fins. Version 1.4 and later applies heating to the fins more frequently and this prevents ice from building up. Older versions defrosted less frequently and when the defrost did come on, a lot of water could be produced. Of course the more humid the lab air, the more ice will build up.
You can check the ICOP version by going to the Windows Control Panel and looking at the programs section. If your Waters Sample Manager ICOP version is less than 1.4, contact your local Waters support and they can help you.
Ken Blakeslee
0 -
Thanks a ton...I will check. I think I'm 1.30 so hopefully that's the problem
0