Ammonium Formate mobile phase
Best Answer
-
Hi Mike,
In our lab we use this:
We use formic acid to adjust pH (ammonium hydroxide if your pH dips too low), though we haven't gone as low as 3.0.
If you have to use more than 2 mL of formic acid to adjust 1 liter of mobile phase, I might be a bit nervous to run this into the TQD - your sensitivity could start to drop due to the amount of formic acid. I've heard of some assays using up to 0.5% formic acid mobile phases, though.
Good luck,
Laura
0
Answers
-
Ammonium Formate is routinely used for Mass Spectrometry applications.
My recommendation for mobile phase reagents and additives is and has always been to obtain HPLC grade or LC/MS grade if available. Yes, Fisher Optima is not inexpensive, but neither are the problems associated with mobile phase impurities.
Whatever you decide, make sure you store it properly in a dessicator as ammonium formate is very hygroscopic.
I have attached some documents that discuss LCMS contamination, mobile phase purity and common background interferences.
I hope you find them useful.