Pressure problem with UPLC Acquity BSM
Answers
-
HI, I don't have much experience with waters instruments yet.. but with the Hitachi intruments I would repair when the pressure would fluctuate as shown I would either change or sonicate the flow valves before and after the pump head. hope this solves your problems.
0 -
Hi
This looks as either an inlet check valve problem on the primary B pump head or a seal leaking on the same head. Please observe that you should not swap a primary pump head with an accumulator pump head since the check valves on these pump heads are not the same.
I would start by changing the check valve. If you do not have a spare one take it off and put it in a sonication bath with first 100% water and the 100% organic for 20 minutes each.
best regards
Rune
0 -
Thanks a lot for your answers.
I have delete the possibility that the problem is the check valve because I changed the Chek Valve B with Check Valve A and the problem rested on B and no on A.
When I changed the pump head from primary to accumulator I changed too the check valve. I changed only pump head.
Best regards
0 -
Hi Again
I would then try the accumulator check valve on the B pump.
best regards