Autosampler leak
Best Answer
-
Hi De Leon
I think what you are seeing is the wash solutions coming out of the wash tube and not a leak. Are you seeing this after the injection valve has turned to inject ? If yes this is normal. When the needle is plunged into the wash station ( in PLUNO this happens after the injection) the weak wash solution is pumped up from the bottom of the wash station and the strong wash solution is pumped out through the needle and meets the weak wash. The combined solution then flows up on the outside of the needle and flows out of the wash tube at the top and then is going to the drain in the wash block. I would instead look for leaks at the injection valve itself or at the column inlet/outlet.
You could measure the flow after the detector to see if you get the flow rate that you have set on the pump.
best regards
Rune
0
Answers
-
Thanks a lot for your answer!
Best regards, De Leon