loop offline or online?
Comments
-
Hi Geatan
Reg. Question 1: Yes the loop stays on-line with the pump for the entire runtime.
Reg. Question 2: The loop off-line time should be long enough to enable the sample to be flushed from the loop onto the column. I usually say 10 times the loop volume, but if you have sticky samples you might need to wait longer. The next sample will not be loaded until approximately 30 sec. before the end of the run anyway (depending on injection volume and loop size).
Reg. Question 3: In PLUNO mode the injection valve cannot switch to off-line before the run has finished. In this mode the sample is drawn to the other side of the injection valve BEFORE the valve turns to load (loop off-line) position. So in load ahead mode the next sample is drawn to the other side of the injection valve during the previous run, and when the run is finished the injection valve turns to load and the sample is pushed into the loop and the injection valve switches back to inject.
I have attached some graphics that might help explain what happens.
best regards
Rune
0 -
Thanks Rune, it is all clear now.
Best regards,
Gaetan