Requirement for using a 5890 GC on Empower 2?
Best Answer
-
Hello
As you have noted the 5890 GC usually contain one of these serial cards:
- HP 3365 ChemStation Interface HP-IB/RS232, part number 19257A
- Serial RS-232-C interface board, part number 19254A
I am assuming yours is the 19257A.
Empower does not use IEEE/GPIB to control the 5890.
Only 5890 GCs with part number 19254A are compatible with Empower software. If your 5890 GC contains part number 19257A, the options are to :
- Purchase a new 19254A board (part number 19242-60030) and RS-232 cable (partnumber 19242-60500) from Agilent
- Connect the 5890 GC to Empower using a Sat/in module.
- Connect the 5890 GC to Empower using the Avocent 8-port Serial Hub.
- Connect the 5890 GC to Empower using the Equinox Serial card.
We support the Equinox card for Windows XP operating system only. The Avocent 8-port Serial Hub is available for both Windows XP and Windows 7 in Empower 2 and Empower 3. Or you can connect to the buslac/e card using the RS-232 with ppropriate adapters and cables.
There are settings that will need configuring on the 5890 GC itself for proper communications.
Have you contacted your local Waters Service?
I hope this information is helpful.
Dot
0
Answers
-
Thanks for the info Dot, found the document I needed googling the part numbers you listed.
http://www.waters.com/webassets/cms/support/docs/71500044403ra.pdf
Didnt contact service personel yet, I hoped there was FAQ already somewhere on the Waters database.
Many thanks