y-axis for chromatogram display
Comments
-
AU stands for absorbance units and is related to transmittance via:
AU = 2 - log10(%T)
AU and height (uV) are directly correlated from the initial zeroing of the absorbance (Ie 200,000 Height = 0.2 AU). However, as your baseline drifts you'll notice these values will not correlate as height is calculated from the baseline that you draw and the y-axis obviously does not change. For this reason, your area will also not correlate with your axises directly most of the time. Area uses a calculus function called integration to calculate the area under the curve in units of AUTime. If the baseline surrounding your peak is identical to the starting AU value, I believe you should find a correlation between uV*sec and AU*min.
I don't think you can change the y-axis so that it correlates to area or height for the above reasons. You can rename your axis but they are only labels and will be ignored in calculations.
0 -
So we can still convert out Height (uV) to Height (AU) by dividing the uV value by 1,000,000. It just won't line up exactly with what is seen on the chromatogram display
because of the baseline offset. Is that correct?
Thanks!
0 -
I believe so yes. When you draw a baseline that has a different AU value from your initial baseline's AU (probably 0, as it's usually set to zero before injection), the correlation between height and AU is lost. You can keep the correlation if you incorporate the difference between your current baseline and your initial baseline.