Failed to obtain Waters GI pump status information, Waters 2525 pump
Best Answer
-
Hello,
The following are some FAQs that might be helpful to review:
· When loading the ICOP for 2525 pump for control with MassLynx, you will need to only load the “Waters Pump Control Software version 1.12.1218.
· Do not load any other ICOP as this will cause system faults.
· BGM and QGM pumps cannot run on the same system.
· Supported Waters software: MassLynx v4.1 SCN 562, 627, or 714
MassLynx PC requirements:
· Appropriate PC configuration (described in the Empower or MassLynx release notes) and Windows® XP Professional or Windows Vista™, with two network cards (10/100/1000 Mbit/s Ethernet).
· Standalone mode PC requirements: Minimum of Intel® Pentium® 4, 3GHz processor, 512 MB memory, 40 GB hard drive, Windows XP Professional or Windows Vista, with two network cards (10/100/1000 Mbit/s ...
0
Answers
-
Thanks for the info Dot!
In addition to the ICOP software the firmware was also found to be an issue by the Waters service engineer. I had a battery replaced and the firmware was flashed to the latest revison 1.14. This firmware is a generic driver that is actually imcompatable with the 2525 pump, so I flashed it back to 1.10 and it worked again :-)
0 -
Excellent! thank you for update.
0