Failed Dynamic Leak test
Comments
-
Darren,
The pressures that you describe for the primary pump heads are normal. The pressure that you set on the diagnostic page for the leak test is used for the Accumulator pump head (the one on the right that is tested first). The software then sets a lower pressure for the primary pump head. This is to ensure that the Accumulator check valve is not forced open during testing of the primary head (so that it is isolated for the test.) Further, if the utility can not reach the setpoint, you get a different failure message, "Unable to compress...
As for your sensitivity issue, although not impossible, the pump is generally not the cause of sensitivity issues. The common mode of failure for the pump is eratic or drifting retention times. Are the retention times for the peaks of interest correct?
Marc
0 -
Hi,
Did the pump pass the static leak tests?
Regards,
Michael