Sensitivity issue: dissappointed from high-sensitivity cell
Comments
-
Hello
Can you share the mobile phase composition and the detector settings
many TX
Liz
0 -
Hello Liz,
mobile phase is Acetonitrile / Water 90:10
The detector settings are:
General:
Sampling rate: 40 points/sec
Filter Time Constant: Normal
Exposure Time: Auto
UV blocking filter disabled
2D Channels (Channel 1):
Data Mode: Absorbance
Wavelength: 290
Resolution: 4.8 nm
The Merck-Hitachi detector is a L-4250 (with no other settings than the wavelength)
Thank you
Florian
0 -
An answer from a Waters RDE scientist follows:
The peak absorbance is scaled by a factor of 5x, which is the path length ratio, and the wavelength is 290 nm where the mobile phase is not likely to be an absorber and a primary noise source.
This looks like an isocratic HPLC separation with a sampling rate of 1 Hz and a 1 second time constant. We recommend that we match the sampling rate and use a "normal" time constant. We suspect that what we'll see is a dramatic improvement in the noise. The HPLC data appears to be captured on a stand alone integrator which implies classical HPLC conditions.
poated by Pat Young (Waters product management)