Sample Set stops early - Empower 2 / Agilent 1100 stack
Answers
-
Hello Jeff,
You did not state the instrument control software version installed. However, based on the version of Empower 2 and service pack I am going to assume it is ICS 1.06.
If there is an Agilent 1100 DAD in the system, then a shutdown method specified with an A1100 DAD system, might encounter an instrument failure at the end of the sample set. To avoid this problem, add a zero volume injection line to the end of the sample set. This line should contain a method set with an instrument method to shutdown the system. The method on this line will run for the specified run time. The run time should therefore be set long enough to allow the shutdown changes to occur.
The reason for not incorporating the shutdown parameters directly in the instrument method that is routinely using on each injection line is when using an A1100 system, the instrument method stops running at the end of the run time and the system reverts back to initial conditions. (Conversely, when using a system containing Waters’ equipment, the instrument method continues to execute after the specified run time, until another injection occurs.)
How many systems are on the LAC/E32? It looks like five (5). It may be worth removing the busSAT/IN system, not just disabling, to see if this is interfering with it for some odd reason.
Let me know if the suggestions are helpful.
Dot
0 -
Dot,
The Agilent LC Control Software version is 1.05.1094. Is this the version we should have? Is there an update that should be applied?
There is a DAD on the system in question. However, we also have another instrument stack that includes a DAD, and to the best of my knowledge this behavior has never been seen with that system.
I'm not totally clear on your comments with regard to the shutdown method. We typically have a separate instrument method with reduced flow, column heater off, etc., and choose that in the "Shutdown Method" pulldown prior to starting the Sample Set. So, the shutdown conditions are not part of the primary instrument method being used - they are separate. Should this approach work properly, or are you suggesting that it's better to not use this function in this way?
There are 4 instruments configured on the LAC/E32, but 5 connections to it because the busSAT/IN module has its own output. So, 4 LAN connections plus a 9-pin serial connection to the COM port for the busSAT/IN. We can certainly try removing that, but the problem that we're experiencing with the insturment failure is intermittent - it's not every run, or every week, it's just occasionally. So if we unplug the busSAT/IN and don't see the problem, that doesn't really mean anything, and we can't leave it unplugged indefinitely since we use the ELSD sometimes.
Thanks for your suggestions, and I look forward to any other thoughts you may have.
Jeff
0 -
Hello Jeff,
The current version for the Agilent LC is ICS v1.06.
I have provided a link to the record, DWNL10090750, on the Waters website. You will be able to download the software, release notes and the certificate of structural integrity. If you choose to order the Waters part number is 667003705.
http://www.waters.com/waters/support.htm?lid=10090750&cid=511442&type=DWNL
The ICS should be installed on the acquisition node where the instrument is attached, Citrix server (if applicable), and any nodes that have the full software stack for Empower client.
The procedure you are using for shutdown is good. I suggest you not change it at this time.
When the behavior is intermittent with one system and other systems do not exhibit the same symptoms, we need to look for differences. Perhaps comparing the firmware revisions or other setup differences should be explored. You pointed out the busSAT/IN module. Look for any other diffierences.
Let me know if this is helpful.
Dot