Plate count

Options
<p>Hi,</p><p></p><p>I have a few questions on plate count. Any input is appreciated.</p><p>I am using an Acquity UPLC BEH 1.7µ, phenyl 2.1 x 150mm column and I am seeing theoretical plate counts (USP) around 70,000. I am more familiar with running HPLC methods and seeing theoretical plate counts up around 2,000. How high could a plate count be on UPLC, and more specifically how high for this column?</p><p>I seen a post on another forum with this question: "What plate count is safe to use to comply with FDA regulation? USP? EP?" -These are basically just different calculations for the same thing. Could they be considered to be equivalent? Using the column mentioned above I am seeing plate counts values around 70,000 using USP calculations and 135,000 using EP calculations for the same peak in the same sample. Has anyone any input on this?</p><p></p><p>Thanks</p>

Comments

  • rune
    Options

    Hi Scurran

    For a well-packed column with fully porous particles (like the UPLC BEH Phenyl 1.7um) the plate count can be estimated to N = 500 x L/dp, where L is the column length (in mm) and dp is the particle diameter (in um).

    This will give around 44000 plates on your column, so the 70000 you get sounds a bit high. Is this a gradient separation ?

    best regards

    Rune

  • Yes, this is a gradient separation.

  • rune
    Options

    Hi Scurran

    I suspected this. According to Snyder and Dolan in their book "High-Performance Gradient Elution", "an experimental value of N in a gradient elution cannot be calculated by the usual expression  for isocratic separation". If the equation is used it will result in much too large values in N.

    Regarding the question whether to use the USP or EP calculation of plate count : The two different methods use the width at different points on the peak. The EP uses the width at half height and the USP use the width where the tangents cross the baseline. If the peak has a perfect Gaussian peak shape (symmetrical) the two calculations should give the same value, but the USP will be more sensitive to tailing, which is why you typically see lower values than the EP.

    In my opinion plate count is mostly usefull to monitor performance/degradation of the column with age/use, and not to say something of the quality of the separation. For this the resolution between all peak pairs is more relevant.

    best regards

    Rune

    P.S.: In the UPLC console you can see what plate count the BEH phenyl column gave when it was tested at the factory. Look under column>column QC.

  • Thank you for this information.