Hi Kate
Is your server and LACE on a different time zone?
Thanks
Si
0 -
Hi Si
apparently it was. We later checked the times on LACE nad it was 5 hous behind. So when it lost connection to the server it got the tame stamp form the LACE. Whn it was reconnectedc it again got the timwe form the server. The LACE is still under observation, because IT are convinced they changed it to GMT, which is the time zone for us. So it shouldn't be really reading american time.
Kate
0 -
Hi Kate
Perhaps someone from Waters can comment on this but I seem to recall when moving to Empower 2 years ago, we set our LACEs to GMT and the server to EST (as it is based in the US). This caused us problems with Agilent 1100s. It can cause an instrument failure if the time difference between client and server is greater than (I think) 1 hour. Not sure if this has been sorted in recent ICOP/Feature Release updates but we still do this today as it is embedded within our SOPs.
Thanks
Si
0 -
That's interesting
Thank you
0 -
Hello Kate,
The responses have all been excellent and good suggestions but maybe we can offer some things to check.
There is a technical note that describes how timestamps are assigned. It is TECN1852729 “Origin of date and timestamps in a client/server environment”. I have provided a link to the document for you so you do not need to search.
http://www.waters.com/waters/support.htm?lid=1852729&type=TECN
This might be helpful in understanding the timestamps and their creation.
After reading the description you have provided here a few suggestions that might be helpful.
- Connect to the Lac/e32 and remotely login, go to the Control Panel/Regional and Language Options applet. Select the Regional Options tab and insure the settings are correct as well as the time.
- Check the recover.log file on the Lac/e32 at or before the date/time of the occurrence. It is possible that the Lac/e32 went into buffering mode if the database was not available and then reconnected to the database. This would be normal. Technical note TECH10007994 “Factors that may contribute to frequent buffering” writes about buffering in an environment. http://www.waters.com/waters/support.htm?lid=10007994&type=TECN
- Technical notes refer to a Time Zone update. You might have already performed this update in your environment. However, if not perhaps, they would be good to review to see if they are applicable. TECN10008076 “Empower Build 1154 and Empower 2 Build 2154 support for Daylight Savings Time”http://www.waters.com/waters/support.htm?lid=10008076&type=TECN TECN1851664 “Daylight Saving Time changes and Empower Build 1154” http://www.waters.com/waters/support.htm?lid=1851664&type=TECN TECN1852458 “Testing complete for Daylight Saving Time Hotfix from Microsoft” http://www.waters.com/waters/support.htm?lid=1852458&type=TECN .
Before making any changes or applying updates check your company SOP and insure you have a good backup.
I hope this is helpful.
Dot
0 - Connect to the Lac/e32 and remotely login, go to the Control Panel/Regional and Language Options applet. Select the Regional Options tab and insure the settings are correct as well as the time.
-
Thanks Dot,
very helpful
Cheers
Kate