How do I analyze an amine by UPLC/ESI/TOF?
Answers
-
Hello
Do you have an HPLC method as a starting place?
Thank you!
Liz
0 -
Yes, I do have an existing isocratic HPLC method which uses a differential
refractometer (Waters' 410) for detection and two 7.8 * 300mm GPC columns
in series (Ultrahydragel 250 and 120) for separation.
I am currently working on a way to analyze the tertiary alcohol amine via
UPLC-ELSD using a C18 2.1 * 50mm. I am very new to using a UPLC-ELSD as
well as developing applications for analyzing compounds using this
particular setup. So I am truly using this as a way to help familarize
myself with conducting method development for LC.
Thank you for your reponse.
?
Sherry
0 -
Sherry:
I sent this off earlier to one of our Chemical Analysis Applications Specialist and this was her suggestion, take generic mobile phases: H2O+0.1% FA, MeOH+0.1% FA - she says she is assuming the amine will ionise by protonation to form a +ve ion so FA will be fine to use.Take a 2.1 x 5 mm BEH C18 column and use a generic screening gradient - James's gradient for ToF screening would be ideal (it doesn't matter that the MPs are different). Prep a 1 ppm solution of the amine alone (pref in H2O - or MeOH @ 1 mg/mL and dilute w H2O). Inject this solution and run the generic screening gradient.Find the monoisotopic exact mass of the amine to 4-d.p. and extract that mass from the acquired TIC. The peak you then see is the amine - hey presto...... analysis of an amine using ToF MS :-)
Hope this is helpful,
Liz
0