System run dry - pump issues?
Answers
-
Hi Timmy
What solvents did you have on the system when the pump ran dry ?
best regards
Rune
0 -
Hi Rune,
Line A is 100% buffer, and line B 100% ACN.
I am anxious to clean line A out obviously but am unsure of what progress is being made as it is not reaching waste? Do you have any suggestions?
Thank you very much for your help
0 -
Hi Again
My guess is that the inlet check valve on the primary pump head is blocked by precipitated buffer, and maybe also the chek valve on the accumulator pump head. I assume that you are trying to prime with pure water, but that the buffer will not dissolve since you are not getting flow through the pump.
It could also be the filter that is mounted before the check valve on the primary pump head. You can easily check this by disconecting the filter and check if solvent is the flowing freely trough the filter. If this is the case you should move to the next step :
If you have a spare inlet check valves I would change this. Try the primary pump head first and see if this helps, if not move to the accumulator pump head. If you do not have any spare check valves you could try and remove them and sonicate in water for ½ hour, to see if you can dissolve the buffer.
Good luck.
best regards
Rune
0 -
To avoid prolonged operation of the pump when solvent reservoirs have run dry consider using a Low Pressure limit. Enter a value significantly lower than your operating pressure to avoid inadvertent shutdowns. For example, if I were doing gradient work and my typical operating pressure range spanned 4000 - 8000 psi, I'd feel comfortable setting the Low Pressure Limit to 2000.
The attached screen shot (from MassLynx) shows the location of this parameter.
Regards,
Rich