Contaminants m/z 113.996 127.9
Answers
-
Elia,
I looked up the masses you are seeing in a table of common contaminants and adducts seen in mass spectrometry and could not identify these two masses with any certainty. In positive electrospray, it is unlikely you are seeing TFA contamination. This would be more easily seen in ES- mode. However, the good news is that you definitely see a reduction in these peaks after system cleaning. I was wondering what your cleaning procedures were? Perhaps you simply need to flush the system longer and with a variety of solvents. First, we recommend high quality (Optima Grade) solvents and additives if you can get them. I have also attached a thorough cleaning procedure that should remove contamination associated with the system. Do not allow any of these solutions to go into the mass spectrometer. Simply disconnect it. You may also want to clean the sample cone thoroughly to rule out MS system contamination.
Finally, have you replaced both the guard column and analytical column? Do these peaks appear when both of these are replaced?
Please keep us posted on your progress.
Ken
0 -
Hello:
It could also be mchanical and that the pump is delivering inconsistently causing the Formic Acid to elute in fits and starts.
Could you send the primary and accumulator traces and the mobile phase conditions. Can you confirm that you are running isocratically and dialing in the required mixture. If so run with pre-mixed solvents and test to see what happens.
Many TX
Liz