UPLC/TQD injection problem
Comments
-
Mike:
It is interesting to me that you stated "tried another instrument method". I would compare the intrument methods and see what the differences were.
There is a capability in ACQUITY UPLC to create a service profile (under the console troubleshooting menu) and this includes a listing of all the log files, which we could inspect, for MassLynx, or for Empower, you can look at the console logs directly. Are you using Empower or MassLynx to control?
I suspect a syringe speed setting and/or priming issue - which I think you supect this too.
Let us know,
Liz
0 -
We are using MassLynx. What do I do with the service profile once I create it?
thx
Mike
0 -
Mike
Can you look for the log file (Program Files/Waters Instruments/LogDatabase) adn take a look at the list of messages from when you were working, does anything jump out.
Many TX
Liz
0 -
Hi Liz
The only alarm error received is "Zp-axis h/w needle lost steps (520)." This was when starting a run right after I did needle seal test (which passed).
The main difference between methods is the Psycho method I've been working on is a gradient vs isocratic (same mobile phase), ESI+ vs ESI-, and a different column is used. There were no pressure issues though and I was getting spray at the probe.
Mike
0 -
Mike
Then that does not sound like an instrument issue, it sounds like smething else. Any other suspicious occurances since then?
Liz
0 -
Occasionally I get the error "Degasser communications HW error." Last happened on 4-15-10. Have to reset the BSM. Seems to be ok after resetting.
Mike
0 -
Hello,
I spoke to Global Service Support and they said that when this error is typically generated we suspect a failure at the electronics level. This could be caused by loose cable connections or related to the circuit board.
So depending on your level of comfort - this can be left to a Waters Service Engineer to check, the normal recommendation would be to turn off power to the BSM and unplug the instrument from the power source. Remove the right side panel to the instrument. Reseat cables on the degasser circuit board and check cables at Personality PCB A (QSPI Interface). If problem persists the issue could be related to a bad Degasser PCB.
Let us know,
Regards
Liz
0 -
Thanks Liz, I called service and they said to keep an eye on it and they'll send someone if or when it happens more consistantly. I'm not comfortable taking off the cover and resetting the cables myself.
Mike
0