Lauren,
In many cases we find that a setting in the range of 0.05-0.1 works better as a starting point than the default 0.5. This would be a welcome change.
Thanks for asking,
Kevin Kolodsick
Consultant Chemist
Analytical Sciences R&D
Eli Lilly and Company
0 -
Lauren,
A survey amongst ourselves reached a consensus of 0.00. We usually start from there and work our way up.
Tony Pecora
Pfizer Consumer Healthcare
0 -
Hi Lauren,
I always start with a touchdown percent of 0.1% and this seems to work pretty well for most of our methods so anything in the region of 0.05 - 0.1% would be a welcome change I think also.
0 -
I also start low and work my way up. If I have one particularly troublesome peak, I'll leave the %TD high and bump up the PW for the odd peak if it comes out too narrow w/ the PW used elsewhere.
0 -
Lauren,
In our laboratory, we always start with a setting of 0.05 % and in many cases it is still too high.
Damien Brunner
BASF Pharma
0 -
As you are talking about future versions of Empower, could it be an option set during project initiation? Maybe where you allow Apex tracking or not, also have a default value for touchdown and liftoff? You should know about what to expect for existing methods.
0 -
I agree that it would be nice for each installation to be able to set their own default value, either in the System Policy/Data Processing Policies where you choose the default integration technique, or as part of the New Project wizard, or both.
Also - thanks for soliciting input from the community!
0 -
We typically use values from 0.05 to 0.01. Defaulting the value to 0.000 would be consistent with the % Liftoff default value. I suggest defaulting it to 0.000.