Split peaks
Answers
-
Hello,
In case you are interested, there was a previous posting earlier this year that was almost identical to yours:
http://forums.waters.com/clearspace_community/thread/1354?tstart=0
Could you please clarify something for me? You mention that you are experiencing problems with XBridge columns BUT you are also using XBridge columns without issue. I would appreciate it if you could clear that up for me.
Thank you very much,
--Doug
0 -
Hi Doug
Thanks for your response. I have seen the problem with all the XBridge columns I have used with this buffer at pH 10; keeping it at 9.5 with methanol (which I think lowers the pH slightly) is fine. It may be conincidental but as this has happened with a few columns now (3 or 4), including a Phase II Acquity column, which makes me think in some way it must be related to the pH. Perhaps, as you say in the other case linked above, it is a conbination of factors, pH, temperature and pressure, but I am concerned about trying it again. The user in the similar case says a Van Guard column helped - how would this help?
Many thanks,
Nicola
0 -
Hi Nicola,
Tracking down such things down can be difficult. Have you confirmed that all of your split peak columns are voided? As you probably know, split peaks can be caused by a contaminated frit/column as well. We've seen this lots of times.
Assuming they are voided, this can be caused by either a mechanical void (bed shifted) or chemical void (bed dissolved) or a combination. It's tough to determine exactly what is the cause. One possible way is to run the same method except use low pH (e.g., 0.1% HCOOH). Under these conditions the packing material will surely not dissolve. If the column lasts, we know that pH is playing a (bad) role. If the column still voids, you start considering a mechanical (pressure) cause. Unfortunately, I do not know of any other way to figure this out. Even autopsying a voided column cannot distinguish between a mechanically vs. a chemically voided LC column.
To make matters more confusing, as you've read and probably know, the BEH material is quite pH resistant.
Not sure if this helps or not. If there is anything else I can do, please let me (and the Community) know.
Best regards,
--Doug