Rezeroing my PDA
Answers
-
Hi,
Empower 2 has an autozero timed event available for the Aquity PDA in the instrument method/events tab.
0 -
Thanks, Mike. I'm using MassLynx, but it sounds like that function is there too.
Doug
0 -
You can turn off the power of PDA , then turn on it again . Detector can be initialized automatically . you can try this means!
0 -
There is a way to set your MassLynx scale to read negative dips and I think that will solve the problem. I am going to check exactly how and get back to you asap. Please bear with me.
Liz
0 -
From the chromatogram view
Display
View
Normilize Data to: Lowest point
0 -
Here is a full explanation...
If this you are using Masslynx and an ACQUITY PDA, he can reset the PDA baseline shown from the Chromatogram Display menu. The default view is to Normalize the display to the largest peak found AND the baseline at zero. Since a Mass Spec does not have to deal with masses below zero, the default is aimed at the MS detectors. Under the Chromatogram window for the detector display that is of interest (highlighted if multiple channels or functions are seen), choose the top Display menu, then the View selection in the drop down menu, and choose the baseline to be set to 'Lowest Point' in the View window that opens. Under most separation buffer and gradient conditions, this 'Lowest Point" displays the potential drifts of individual wavelengths or Total Absorbance Chromatograms.
Hope this helps,
Liz