False positive results
Answers
-
Hello
This is interesting, so for the same vial (sample 2) only the first injection showed a high response? After this they were reproducible? The standards that bracketed them were reproducible?
Could we see a chromatogram?
What is the sample format - vials, plate etc?
Could you re-inject with a blank in between?
Many TX
Liz
0 -
Liz,
Below are the answers to your questions.
This is interesting, so for the same vial (sample 2) only the first injection showed a high response? yes
After this they were reproducible? yes
The standards that bracketed them were reproducible? yes
Could we see a chromatogram? It is attached. It is sample 75 and not sample 2.
What is the sample format - vials, plate etc? vials
Could you re-inject with a blank in between? I will be injecting water sample extracts tonight and can include blanks.
Thanks,
MopedGirl
0 -
Many thanks
Liz
0 -
Is it right that you use only water as sample solution?
I experienced (but not very often) problems if a compound is drawn throught a peek capillary (like the uplc standard needle) and no organic solvent is used in sample solution. Even the solubility of the compounds are higher than the concentration there is always the possibility that matrix or metall ions or whatever else leads to adsorbtion effects on the surface of the peek capillary for the compounds AND affected later injections when the adsorption effect comes to an end - most likely it affects only one injection.
We get this problem solved by replacing 5% to 20% of the water used for sample preparation with an (or adding 10% of) organic solvent.
One other possibility is you experienced is a hot spot within your sample. And if you run into the hot spot problem, you have to check your whole sample preparation if there is a step where hot spots could be build or contaminate your sample. Because hot spots can have a wide range of reasons there is no overall solution.