oxycodone/oxymorphone separation
Answers
-
these compounds seem to be prone to fronting
0 -
You may have success with a BEH Phenyl column. Mobile Phase A: 5mM ammonium bicarbonate pH 10.5 (use NH4OH to pH)/MeOH 90/10 Mobile Phase B: Acetonitrile Column Temp 50 deg C flow rate 0.7ml/min
Isocratic or Initial conditions for a gradient A:B 70:30
Tailing for Oxycodone should be 1.1 - 1.2
0 -
Great advice Kenneth. I will certainly try your methodology. I wouldn't be opposed to hearing a promising HILIC methodology either.
0 -
Hi,
You may try,
Column : BEH HILIC Column, 2.1 x 100 mm, 1.7 µm
Mobile : 85:15 {0.1% ammonium acetate in acetonitrile:0.1% ammonium acetate in water, pH 4.2}
WL: 254
Column temp. : 40C
Flow: 0.5
0 -
Thanks for the great advice. Are either of these recommendations currently being utilized for oxycodone/oxymorphone analysis?
0 -
try:
10mM orthophosphate buffer at pH 8.30
BEH C18 30 x 2.1 mm,
Acetonitrile ramping from 5 to 50% over 1.1 minutes at a flow rate of 1.5 mL/min, remainder buffer.
Temperature = 50oC
Oxymorphone RT = 0.5 min, Oxycodone RT = 0.8 minutes
enjoy!
0 -
What kind of resolution are you getting with that setup?
0 -
If you're fronting, you're likely overloading your column. Dial down the injection volume or the concentration of the finished sample, or a combination of both.