PDA Peak purity How to process data?
Answers
-
Hi,
This procedure was developed on M32. I suppose you have Empower (or higher versions of Milliennium). There Auto treshold is typically used. If Autotreshold returns standards (which should be pure) to be unpure (purity flag...) then I would go for alternative method, like the one described in the Dr. Levins paper. But you will have the problem that you can not set up more than one solvent angle. One possibility is to put in the highest purity angle you get from 6 peaks from 6 iniections and key it in as solvent angle. Another option is also to make six injections and make a library match: one peak is saved in library, the others are compared and the highest match angle can be to my opinion set as solvent angle. I would also like to say that I would increase solvent angle calculated from either methods to cover some of the uncertainty.