PG analysis by UPLC/MS/MS
Answers
-
Hello Mike,
I spoke with Peter Lee, a Principle Applications Scientist in Milford, and he has the following comment. I hope it is helpful.
"The abundances of the fragment ions are governed by the steric configuration of phosphatidylglycerol (PG). For example the regioisomers,16:0/18:1-PG (m/z 747) and 18:1/16:0-PG (m/z 747), have very different of ion ratio. According to the data reported in J. Am. Soc. Mass Spectrom. 12 (2001), 1036-1043, the ion ratio of 255/281 is about 0.5 for 16:0/18:1-PG and about 2 for 18:1/16:0-PG.
According to the information provided, I think perhaps the specimen and standards have different regioisomer ratios. I think he can determine isomeric differentiation of specimen by creating an ion ratio calibration curve with isomeric pure standards." - Peter Lee
Ken Blakeslee
0 -
Thanks for the reply, I'm going to try and get that article.
Unfortunately it doesn't explain my situation as while I am getting about 0.59 IR for 16:0/18:1 PG (Avanti pure standard), I'm getting 0.38 IR using egg yoke PG. And Avanti doesn't make 18:1/16:0 PG so I can't compare the two.
I'll get the article and see if I can glean something from that.
regards
Mike
0 -
Peter Lee responds:
The 0.38-IR of 16:0/18:1 PG (from egg yoke PG) is very similar to the reported result (Figure 1a in the attachment). According to the theory reported in J. Am. Soc. Mass Spectrom. 12 (2001), 1036-1043, the samples having high IR might contain some regioisomer 18:1/16:0-PG. It is possible to estimate the ratio of 18:1/16:0-PG in the samples using the data in Figure 1a and 1b if the regioisomers 16:0/18:1-PG and 18:1/16:0-PG are co-eluted.
0 -
Thank-you very much kind sir .
And thank-you for posting the article.
Regards
Mike