Polyetheramine Analysis Using UPLC with SQD
Answers
-
I have emailed your request to SQD operator's at Waters. However, as many are on vacation it may be a delayed response.
Liz
0 -
Apologies for the delay and I was advised that this should translate easily to SQD. The desolvation gas and temp may need to be increased a bit but should be fine on SQD
Happy New Year!
0 -
Thanks for the information. It will help in establishing a new method.
0 -
Thank you Liz for your answers, just a little additional information.
As a starting point for method development, I would suggest using the following parameters for the SQ Detector when using the ACQUITY method from the apps note you originally attached;
Desolvation Gas Flow; 900L/Hr
Desolvation Temperature; 400°C
Source Temperature: 150°C
These should be good generic starting points for LC methods with flow rates above 0.4mL/min.
Thanks and all the best,
Ed0