Custom field with internal standard
(ARstd2/ARstd1)*(Wstd1/Wstd2)*100
ARstd2=Area(solvent)/Area(internal std) (of second standard)
ARstd1=Area(solvent)/Area(internal std) (of first standard)
Wstd1= weight of first standard for each solvent
Wstd2= weight of second standard for each solvent
How can i make it for each solvent. Do you have any idea?
Comments
-
For solvent I presume you mean peak? And in your formula, for the Area(Internal standard) of first standard is that the average area of internal standard over the 5 injections?
0 -
The solvents are 5. Each solvent is an individual peak in one mixture with internal standard inside (another peak). So 6 peaks. 5 injections of a mixture of 5 peaks and one peak of internal STD. The average of each Areapeak/ Areainternal std0
-
Three custom fields will do this. Label your first standard of 5 injections as S1, the second one as S2 and populate CCalRef1 of your processing method with the Internal Standard peak. Put the standard weight for first standard of each solvent peak into CConst1 of processing method and standard weight for second standard for each solvent peak into CConst2 of processing method (so if your standard weight was 101.998mg put 101.998 into CConst1/2).
Percent_Recovery_A
S2.%.(Area)/S2.%.(CCalRef1[Area])
Percent_Recovery_B
S1.%..AVE(Area)/S1.%..AVE(CCalRef1[Area])
Total_Percent_Recovery
(Percent_Recovery_A/Percent_Recovery_B)*(CConst1/CConst2)*100.
Make sure the function "Summarize Custom Fields" is inserted as the last line of the sample set and the "Processing" Column is set to Normal for that row. Process the sample set to generate a result set, Job done. % Recovery for each of the 5 solvent peaks obtained. Worked perfectly for me when I tested it.
0 -
Actually, I just realized since you are using Internal Standard, then the field "Response" automatically calculates the Area of each peak over the area of the peak designated as Internal Standard for each sample so your calculation can all be done in one custom field:
(S2.%.(Response)/S1.%..AVE(Response))*(CConst1/CConst2)*100
0 -
Hi @Empower2019 let me know if you can help. I am trying to do this CF "STD_Area_Ratio", Area of STD/Area of IS. Unfortunately, the results are consistently equal to 1. Do you have any insights into why this might be occurring? Thank you.
0 -
Are the CComp2 and CComp1 peaks assigned correctly in the row with the standard peak?0