Reporting the number of injections for HPLC
Options
<p>I am running Empower 2 FR5 with Agilent 1100/1200 systems. Maybe I am having a brain freeze but I can't figure out a way to report</p><p>the number of injections for a run. I see Injection (which is just a database unique ID number and not pertinent to each run), Injection (which just counts the injections for the row)</p><p>but I don't see Number of Injections or something similar. I want to be able to report out the number of injections for a run and/or colum so we can see the number of injections</p><p>performed on a system or column over time. Does anyone know how to do this?</p><p></p><p>Thanks,</p><p>Chris</p>
0
Answers
-
Not sure if there is a way to do this with a system field...
You could create a sample set table (in your report) and then sum the injection column (watching out for duplicates)
Otherwise, you could use a custom field. Try this:
This will give you the total injections taken from a single vial -
.%..MAX(Injection)
Then
%.%..SUM(EQ(Injection,1)*PreviousEquation)
I did a quick check and it worked for me.
Nate
0